- Joined
- Mar 22, 2014
- Messages
- 5,357
+/- 1.5 = spread of 3 which is somewhat laughable. I suspect they say that just so people can't return the machine saying it's out of spec. I would hope it has much lower variance than that.
ASTM E18-20 is awesome thank you for the pointer. It even has a precision/bias study (Table 4) that tells you what you should expect in terms of (a) expected error when reproducing your own readings ["r_pb"] and (b) expected error when reproducing someone else's readings ["R_pb"].
I also have one of those certifications with my PHR-1 which shows 5 readings and shows it's within 0.5 HRC of the correct number - but how do you know it's real? Unless it's performed by a 3-rd party trained in they can write whatever they want on it. Even if it's not totally fabricated, it's easy to just drop readings that are wonky until you get a few that look nice enough to write down.
It is a factory "allowance"
The customer has to calibrate it when they set it up anyways.
The regulation plate can be adjusted +/- 5 HRC to fine tune upon setup.
You have to make sure you use multiple hardness blocks at different values so that you are not just "block chasing"
It's good to see you are having some skepticism. This is what fuels a deeper thirst for knowledge rather just asking random questions to random people on a forum.
The fear of uncertainty of measurement in ones own hardness readings shows a deeper level of understanding and humility. It is a reason why those of us that care will only use certified hardness blocks and certified indenters, fine tune the dwells, double check the test forces and do more frequent calibrations and log them to watch how the reproducibility changes over time.
Just like how even the sharpest edge on the longest cutting steel can also wear and dull, critical components of the hardness tester also degrade.
Calibration is like watching how the edge of a knife dulls over time with use and when we need to intervene before problems arise.